Dear All,
I'm designing a medical device which generates shockwaves. I'm facing
serious EMI problems. The device generates the shockwaves by
discharging a high voltage capacitor through a Spark Gap, a triggering
transformer is used. The triggering transformer is triggered using a
circuit. The trigger circuit is fully isolated using an isolated supply
and optocouplers.
A seperate microcontroller controls the triggering process by sending
signals to the opto isolated triggering circuit.
Despite all the isolation measures taken, whenever a HV discharge
occurs the microcontroller circuit resets. I look at the supply voltage
of the uC board and see large spikes.
What I am doing wrong? Is it the layout in the HV section?
Please Help...
 
J

Jeroen Belleman

Jan 1, 1970
0
booth said:
Dear All,
I'm designing a medical device which generates shockwaves. I'm facing
serious EMI problems. The device generates the shockwaves by
discharging a high voltage capacitor through a Spark Gap, a triggering
transformer is used. The triggering transformer is triggered using a
circuit. The trigger circuit is fully isolated using an isolated supply
and optocouplers.
A seperate microcontroller controls the triggering process by sending
signals to the opto isolated triggering circuit.
Despite all the isolation measures taken, whenever a HV discharge
occurs the microcontroller circuit resets. I look at the supply voltage
of the uC board and see large spikes.
What I am doing wrong? Is it the layout in the HV section?
Please Help...

Electromagnetic compatibility must be designed in right from the
start. Adding it later is much harder. If you make medical
devices you should be very aware of that. Be especially vigilant
when you are making sparks, because that usually implies the
presence of wideband powerful magnetic and electrical interference.

Anyway, there are three main paths by which interference may mess
up your circuits:

- Magnetic coupling. To reduce this, keep loop areas of closed circuits
as small as possible. Keep bus bars close together. Twist wires. Output
and associated return leads should follow the same path. Keep loops with
low-level sensitive signals away from loops with high-power signals.
If the field is high frequency, a conductive shield can divert or
contain the flux, depending on geometry. Low frequency and constant
fields may be channeled around sensitive spots using soft iron.

- Electrostatic or capacitive coupling. Separate nodes with rapidly
changing voltages from nodes with high-impedance low-level nodes.
Lower victim node impedances, if you can. Keep nodes compact. Put
grounded or guarded shielding between them. Use metal boxes.

- Common impedance coupling. This is what you get when currents from
more than one separate circuit flow through a common wire or trace.
Even solid conductors can have a significant impedance, especially
when one of the circuits carries a large current and the other a
sensitive low-level signal. This is even more true if high frequency
signals are involved. Give every circuit its own return lead. Make
sure you *know* where return currents flow.

Decoupling or bypass capacitors are a way to reduce loop area and
confine RF currents to small loops. Filters on input and output
leads can divert or impede undesirable currents, if signal and
interference bands do not overlap. Coaxial cable works as a shield
for both electrical and magnetic fields at sufficiently high
frequencies. PCB ground planes reduce common impedance coupling
and offer shielding as well. Signal transformers, opto-couplers or
differential signalling can be used to open unavoidable loops at a
well defined place.

I probably just missed a nice consulting job...

Regards,
Jeroen Belleman
 