Maker Pro
Maker Pro

Lock-in Amplifier at a distance from the signal source (ref)

J

Jeepster

Jan 1, 1970
0
I'm trying to use a lock-in amplifier at a big distance from the
signal source (laser communication between two buildings) and hence
cannot use my reference input.

When I punch in the frequency of the transmitter (signal source), the
lock-in signal output keeps oscillating.

Is this because:
The frequency I punched in is not exact? or
The signal phases are not synced? or
The lock-in just cannot do this?

Any help would be much appreciated,
Jim
 
P

Phil Hobbs

Jan 1, 1970
0
Jeepster said:
I'm trying to use a lock-in amplifier at a big distance from the
signal source (laser communication between two buildings) and hence
cannot use my reference input.

When I punch in the frequency of the transmitter (signal source), the
lock-in signal output keeps oscillating.

Is this because:
The frequency I punched in is not exact? or
The signal phases are not synced? or
The lock-in just cannot do this?

Any help would be much appreciated,
Jim

A lock-in is precisely equivalent to a very narrowband filter followed
by a phase-sensitive detector. Your trouble is that the modulation
carrier is getting out of phase with the lock-in reference.

You can do several things. From simplest to hardest:

1. If you have a two-phase lock-in, just switch to "amplitude/phase"
mode and use the amplitude output. This is exactly equivalent to a
narrowband filter followed by an AM detector, which is probably what you
want.

2. Amplify the daylights out of your received signal and put it into
both the signal and reference inputs of the lock-in. If your modulation
is fast enough, and the lock-in doesn't use a phase-frequency detector,
you may get it to lock to the carrier.

3. Build a phase-locked loop with a very narrow loop bandwidth (much
slower than the modulation of the carrier), lock it to the received
carrier, and use that for the lock-in phase reference.

Cheers,

Phil Hobbs
 
Top