Looking for calibration procedures / manual for a Simpson Model 260 7M Mulitmeter.
Thank you Delta Prime,Quick search didn't take long.
But I'm happy you came here first. And it's free to download.
If you have any other questions please do not hesitate to ask.
Free Simpson Manuals - Calibration - Schematics - Training
www.simpson260.com
Could be the test leads themselves or the mechanical adjustment for the pointer (but this would be apparent at 'no signal' so unlikely).I cannot zero out my Rx1 range on my meter
You're very welcome. In addition to the information stated above.Thank you Delta Prime
Thank you again Delta Prime,You're very welcome. In addition to the information stated above.
Follow this procedure.
1) Check (Ohmmeter circuit) inside
your unit for defective resistors, there is no adjustment potentiometer inside specifically for RX1 scale.
2) short out the banana jacks with a jumper, On RX1 scale, your test leads have more resistance than one ohm. If your meter zeros with jumper. Then you must subtract the resistance of your test leeds on every measurement you take on RX1 scale.
3)Check battery terminals for corrosion or loose connections because... On RX1 scale Independent of "brand new batteries installed" more current is used (Sourced) than any other scale on your meter.
![]()